Mechanical Ventilation Basics
Anaesthesia · Critical Care · lean revision notes
Mechanical Ventilation Basics
Mechanical ventilation is the cornerstone of critical-care and intra-operative support, replacing or augmenting the work of breathing while protecting the lung from iatrogenic injury. For NEET PG, the high-yield zone is modes, lung-protective settings, PEEP physiology, auto-PEEP, permissive hypercapnia and weaning — facts that recur almost every year in Anaesthesia and Medicine.
Definitions & basic terminology
A ventilator delivers a breath defined by three things: what triggers it (initiates inspiration), what limits/targets it (the variable held constant during inspiration), and what cycles it (terminates inspiration and begins expiration).
| Phase variable | Meaning | Common settings |
|---|---|---|
| Trigger | Starts the breath | Time (machine), pressure or flow (patient) |
| Limit/Target | Variable controlled during inspiration | Volume or pressure |
| Cycle | Ends inspiration | Volume, time, flow, pressure |
Key terms you must define crisply:
- Tidal volume (Vt): volume delivered per breath. Lung-protective target = 6 ml/kg of predicted (ideal) body weight.
- Respiratory rate (RR): breaths/min set on the ventilator.
- Minute ventilation (MV) = Vt × RR — the main determinant of CO₂ clearance (PaCO₂ is inversely proportional to alveolar ventilation).
- FiO₂: fraction of inspired oxygen (0.21–1.0); main determinant of oxygenation along with PEEP.
- PEEP (Positive End-Expiratory Pressure): pressure maintained in the airway at end-expiration; recruits collapsed alveoli, raises FRC and improves oxygenation.
- Plateau pressure (Pplat): alveolar distending pressure measured during an inspiratory hold (no-flow state). Reflects alveolar/lung-and-chest-wall compliance. Target < 30 cm H₂O.
- Peak inspiratory pressure (PIP): maximum pressure during inspiration; includes airway resistance + alveolar pressure.
- Driving pressure = Pplat − PEEP. Strong predictor of mortality in ARDS; keep < 15 cm H₂O.
- I:E ratio: normally 1:2; inverse-ratio ventilation (>1:1) improves oxygenation but risks auto-PEEP.
- Compliance = ΔVolume / ΔPressure = Vt / (Pplat − PEEP).
High-yield: Predicted body weight (not actual) sets the tidal volume. Obese patients have normal-sized lungs — using actual weight over-distends them and causes volutrauma.
Classification of modes
Modes are grouped by what variable the ventilator targets (volume vs pressure) and how much patient effort is allowed (controlled, assisted, spontaneous).
| Mode | Control variable | Patient effort | Key point |
|---|---|---|---|
| VCV (Volume-Controlled) | Volume (guaranteed Vt) | Often none/minimal | Constant Vt; pressure varies with compliance — risk of barotrauma |
| PCV (Pressure-Controlled) | Pressure (set Pinsp) | Variable | Constant pressure; Vt varies with compliance — risk of hypoventilation if compliance falls |
| A/C (Assist-Control) | Volume or pressure | Patient can trigger | Every triggered breath gets full support; risk of respiratory alkalosis & breath-stacking |
| SIMV | Volume or pressure | Mandatory + spontaneous breaths | Mandatory breaths synchronised to patient effort; spontaneous breaths between them |
| PSV (Pressure Support) | Pressure | Fully spontaneous | Patient sets RR & Vt; flow-cycled; used for weaning |
| CPAP | Constant airway pressure | Fully spontaneous | No inspiratory support beyond continuous pressure; for oxygenation/OSA |
Volume-controlled vs pressure-controlled — the classic comparison
- VCV: You set Vt and RR; the machine guarantees the volume. Because volume is fixed, airway pressure rises if compliance falls (e.g. ARDS, pneumothorax) — danger of barotrauma. Flow is typically a constant ("square wave") pattern. Best when guaranteed minute ventilation is essential.
- PCV: You set the inspiratory pressure and time; the machine guarantees the pressure. Flow is decelerating, giving a more even alveolar filling and lower peak pressures. Because pressure is fixed, Vt falls if compliance worsens — danger of hypoventilation/hypercapnia. Decelerating flow improves gas distribution and is preferred in stiff (ARDS) lungs.
High-yield: In VCV → volume constant, pressure variable. In PCV → pressure constant, volume variable. This single dichotomy is the most repeated MCQ in ventilation.
SIMV and CPAP
- SIMV (Synchronised Intermittent Mandatory Ventilation): delivers a set number of mandatory breaths synchronised with patient effort; between mandatory breaths the patient breathes spontaneously (often supported by PSV). Historically used for weaning, though pure PSV/spontaneous-breathing trials are now preferred.
- CPAP: a single continuous pressure throughout the respiratory cycle in a spontaneously breathing patient. No tidal cycling support is added. Used in obstructive sleep apnoea, cardiogenic pulmonary oedema, and as a weaning/oxygenation tool. BiPAP = two levels (IPAP > EPAP) and is essentially CPAP + pressure support — first-line non-invasive ventilation for COPD exacerbation with hypercapnia.
Indications for mechanical ventilation
- Hypoxaemic respiratory failure (Type I) — PaO₂ < 60 mm Hg on high FiO₂ (ARDS, pneumonia, pulmonary oedema).
- Hypercapnic respiratory failure (Type II) — rising PaCO₂ with acidosis (COPD, neuromuscular disease, drug overdose).
- Airway protection — GCS ≤ 8, loss of gag, severe secretions.
- Increased work of breathing / fatigue — RR > 35, accessory muscle use.
- Apnoea, peri-operative paralysis, raised ICP requiring controlled PaCO₂.
Lung-protective ventilation (ARDSNet strategy)
The landmark ARMA / ARDSNet trial (NEJM 2000) showed that low-tidal-volume ventilation reduces mortality in ARDS. This is the single most testable "management" fact in ICU ventilation.
Stepwise approach (ARDSNet):
- Set initial Vt = 6 ml/kg predicted body weight (range 4–8).
- Maintain Pplat < 30 cm H₂O → if exceeded, reduce Vt in steps of 1 ml/kg (minimum 4 ml/kg).
- Keep driving pressure (Pplat − PEEP) < 15 cm H₂O.
- Use a PEEP–FiO₂ table to titrate oxygenation; target SpO₂ 88–95% or PaO₂ 55–80 mm Hg.
- Permit permissive hypercapnia — allow PaCO₂ to rise as long as pH stays ≥ 7.20–7.25.
- RR up to 35/min to compensate for the small Vt.
High-yield: ARDSNet trio → Vt 6 ml/kg PBW, Pplat < 30, SpO₂ 88–95%. Low Vt reduces volutrauma and is proven to lower mortality in ARDS.
Severe ARDS adjuncts (PaO₂/FiO₂ < 150): prone positioning (PROSEVA trial — mortality benefit, prone ≥ 16 h/day), neuromuscular blockade (early, short course), high PEEP, and ECMO as rescue (CESAR/EOLIA). Recruitment manoeuvres are used cautiously.
The Berlin definition grades ARDS by PaO₂/FiO₂ ratio on PEEP ≥ 5: mild 200–300, moderate 100–200, severe < 100.
PEEP — physiology and pitfalls
PEEP keeps alveoli open at end-expiration, increasing functional residual capacity (FRC) and recruiting atelectatic lung, thereby improving the shunt fraction and oxygenation. Standard "physiological" PEEP is 5 cm H₂O.
Benefits: ↑ oxygenation, ↓ shunt, prevents atelectrauma (cyclic alveolar collapse/reopening), reduces work of breathing in cardiogenic oedema by lowering LV preload/afterload.
Harms of excessive PEEP:
- ↓ Venous return → ↓ cardiac output and hypotension.
- Barotrauma/volutrauma → pneumothorax.
- Over-distension increasing dead space and ↑ pulmonary vascular resistance (RV strain).
- Falsely elevated CVP/PCWP readings; raised ICP.
High-yield: The chief haemodynamic complication of high PEEP is reduced venous return → hypotension. Treat with fluids and reduce PEEP if needed.
Auto-PEEP (intrinsic PEEP / breath-stacking)
Auto-PEEP is air trapping when the patient cannot fully exhale before the next breath, so gas accumulates and end-expiratory alveolar pressure exceeds the set PEEP.
Causes: high RR, prolonged expiration disorders (COPD, asthma), inadequate expiratory time, inverse-ratio ventilation, high minute ventilation.
Consequences: dynamic hyperinflation → ↓ venous return → hypotension and pulseless electrical activity (PEA) / cardiac arrest; difficulty triggering the ventilator; barotrauma.
Detection: end-expiratory hold manoeuvre; expiratory flow not returning to zero on the flow–time waveform.
Management — flow: Recognise auto-PEEP → disconnect the circuit (allow full exhalation) if haemodynamically unstable → reduce RR, shorten inspiratory time / increase expiratory time (lower I:E), treat bronchospasm, reduce Vt; apply modest extrinsic PEEP (~80% of auto-PEEP) to ease triggering.
High-yield: A ventilated asthmatic who suddenly becomes hypotensive → think auto-PEEP / dynamic hyperinflation (and exclude tension pneumothorax). First step: disconnect from ventilator and allow exhalation.
Permissive hypercapnia
Deliberately accepting an elevated PaCO₂ (and mild respiratory acidosis) to keep tidal volumes and pressures low — protecting the lung in ARDS and severe asthma.
- Acceptable down to pH ≈ 7.20–7.25; below this, increase RR or give sodium bicarbonate cautiously.
- Contraindicated in raised intracranial pressure (CO₂ → cerebral vasodilation → ↑ ICP), and used cautiously in pulmonary hypertension/RV failure (hypercapnia raises PVR).
Ventilator-induced lung injury (VILI)
Four overlapping mechanisms — mnemonic "Bad Vibes Are Bio":
| Type | Mechanism | Prevented by |
|---|---|---|
| Barotrauma | High airway pressures → alveolar rupture, pneumothorax, pneumomediastinum | Pplat < 30 |
| Volutrauma | Over-distension by large Vt | Vt 6 ml/kg PBW |
| Atelectrauma | Repeated collapse–reopening (shear) | Adequate PEEP |
| Biotrauma | Cytokine release → systemic inflammation, MODS | Overall lung-protective strategy |
High-yield: Volutrauma (over-distension by volume), not pressure per se, is now regarded as the principal driver of VILI — hence low-Vt ventilation.
Monitoring & investigations
- Arterial blood gas (ABG): the investigation of choice to titrate ventilation — guides FiO₂/PEEP (oxygenation) and RR/Vt (PaCO₂/pH).
- Capnography (EtCO₂): confirms tube placement, monitors ventilation continuously; sudden loss of EtCO₂ = displaced tube/circuit disconnection/cardiac arrest.
- Pressure–volume and flow–time loops: detect auto-PEEP, over-distension ("beaking"), and patient–ventilator dyssynchrony.
- Chest X-ray / lung ultrasound: tube position and complications (pneumothorax).
Weaning from mechanical ventilation
Weaning begins once the cause of respiratory failure is improving. Readiness is assessed daily.
Readiness (objective) criteria:
- Underlying illness resolving; haemodynamically stable, minimal/no vasopressors.
- Adequate oxygenation: PaO₂/FiO₂ > 150–200, FiO₂ ≤ 0.4, PEEP ≤ 5–8.
- pH > 7.25, intact respiratory drive, adequate cough.
Spontaneous breathing trial (SBT): T-piece or low-level PSV (5–8) / CPAP for 30–120 min — the gold-standard weaning test.
Rapid Shallow Breathing Index (RSBI) = RR / Vt (in litres).
- RSBI < 105 → likely to wean successfully.
- RSBI > 105 → likely to fail (rapid, shallow breathing).
High-yield: RSBI = f/Vt; < 105 predicts successful extubation. This exact cut-off is a perennial MCQ.
Other indices (less used): negative inspiratory force (NIF) more negative than −20 to −30 cm H₂O; vital capacity > 10–15 ml/kg; minute ventilation < 10 L/min.
Weaning flow: Treat cause → daily sedation interruption + readiness screen → SBT (30–120 min) → assess RSBI/tolerance → extubate (ensure cuff-leak, adequate cough, GCS) → consider NIV to prevent re-intubation in high-risk (COPD).
Complications of mechanical ventilation
- Ventilator-associated pneumonia (VAP): pneumonia > 48 h after intubation; prevent with the VAP bundle — head elevation 30–45°, daily sedation breaks + spontaneous breathing trials, DVT and peptic-ulcer prophylaxis, oral chlorhexidine, subglottic suction.
- VILI (as above), barotrauma → tension pneumothorax.
- Auto-PEEP, haemodynamic compromise from positive intrathoracic pressure.
- Oxygen toxicity (prolonged FiO₂ > 0.6 → absorption atelectasis, free-radical injury).
- Tracheal injury, sinusitis, ICU-acquired weakness, delirium, stress ulceration.
Key differentials & decision points
- VCV vs PCV: stable compliance / need guaranteed MV → VCV; stiff lungs, high pressures, paediatrics → PCV.
- NIV vs invasive: COPD exacerbation/cardiogenic oedema with intact airway and consciousness → trial NIV (BiPAP) first; failure or airway compromise → intubate.
- Hypoxia not improving on FiO₂ → it's a shunt → raise PEEP (FiO₂ alone won't fix true shunt).
- Sudden desaturation/high pressure alarm — DOPE: Displacement of tube, Obstruction/secretions, Pneumothorax, Equipment failure.
Recently asked / exam angle
- ARDSNet tidal volume = 6 ml/kg predicted body weight and Pplat < 30 cm H₂O (repeatedly tested as the protective ventilation combo).
- RSBI < 105 predicts successful weaning — direct one-liner MCQ.
- Difference between VCV (volume fixed, pressure varies) and PCV (pressure fixed, volume varies).
- Berlin definition PaO₂/FiO₂ cut-offs for mild/moderate/severe ARDS.
- Auto-PEEP in a ventilated asthmatic/COPD presenting as hypotension; first step = disconnect circuit.
- Driving pressure < 15 as a mortality predictor — newer favourite.
- High PEEP complication → decreased venous return / hypotension.
- Prone positioning improves mortality in severe ARDS (PROSEVA).
- Permissive hypercapnia is contraindicated in raised ICP.
- Decelerating flow waveform is characteristic of pressure-controlled ventilation.
Rapid revision
- MV = Vt × RR; PaCO₂ is inversely proportional to alveolar ventilation.
- VCV → volume constant, pressure variable; PCV → pressure constant, volume variable (decelerating flow).
- Lung-protective ventilation: Vt 6 ml/kg PBW, Pplat < 30, driving pressure < 15, SpO₂ 88–95%.
- PEEP ↑ FRC and recruits alveoli; main harm = ↓ venous return → hypotension.
- Auto-PEEP = air trapping in COPD/asthma; sudden hypotension → disconnect circuit + lower RR + ↑ expiratory time.
- Permissive hypercapnia acceptable to pH ~7.20; contraindicated in raised ICP.
- VILI = Barotrauma, Volutrauma, Atelectrauma, Biotrauma; volutrauma is the key driver.
- RSBI = RR/Vt; < 105 predicts successful weaning; SBT is the gold-standard test.
- ABG is the investigation of choice for titrating the ventilator.
- Berlin ARDS: PaO₂/FiO₂ mild 200–300, moderate 100–200, severe < 100 (on PEEP ≥ 5).
- Prone positioning (≥16 h) lowers mortality in severe ARDS (PROSEVA); ECMO is rescue.
- Sudden ventilator desaturation → DOPE (Displacement, Obstruction, Pneumothorax, Equipment).